Ricoh D191 MPC401SP Z-C2

Troubleshooting. Vsg Adjustment Result

SP3-323-001 to -010 (Vsg Adjustment Result)

No. Result Description Possible Causes/Action
1 O.K Vsg adjustment is correctly done.  
2 ID sensor adjustment error Vsg cannot be adjusted within 4.0 ±0.5V. • Dirty ID sensor (toner, dust, or foreign material)
• Dirty transfer belt
• Scratched image transfer belt
• Defective ID sensor
• Poor connection
• Defective BICU
1. Clean the ID sensor.
2. Check the belt cleaning. Clean or replace the transfer belt.
3. Replace the image transfer belt.
4. Replace the ID sensor.
5. Check the connection.
6. Replace the BICU board.
3 ID sensor output error ID sensor output is more than "Voffset Threshold" (SP3-324-004) • Defective ID sensor
• Poor connection
• Defective BICU
1. Replace the ID sensor.
2. Check the connection.
3. Replace the BICU board.
9 Vsg Adjustment error Vsg adjustment has not been completed. • Other cases
Retry SP3-321-010.